The stereo eyepieces are dark,
It might be that Objective lens is not seated properly or very dirty. So perhaps by rotating the objective block may help before trying to clean it. Also try to start with the lowest magnification to see that you have functional microscope before trying higher magnifications. Good luck.
I'm using the Nikon ECLIPSE
There are a few things that could cause a shadow in the field of view.
First if the condenser is not centered, this can create a crescent shaped shadow and rob your image of light. Center the condenser and raise it to within about a 16th of an inch from the top of the stage.
Second, the slider on the head which diverts light from the eyepieces to the camera must be fully engaged to the "camera" position
Third, make sure that the diaphragm on the condenser is not closed too far. There is an adjustment ring for this. Set it to 70-80% of the numerical aperture of the objective you are using.
Fourth: if there is a "swing out" condenser lens you will need it completely in or out depending on the objective you are using. for 2x-5x you need the swing out condenser lens completely out of the light path. For objectives 10x and up you need the swing out condenser lens completely in and centered in the light path.
If these things are properly adjusted you should not have a shadow in the field of view. If you still have a problem, let me know.
We are using our Nikon Eclipse e200 for the first
If this is a brand new scope then there is most likely some packaging or packaging tape that is preventing movement. Scopes are meticulously packaged because they can be rather delicate instruments. In some rare cases the scope may be damaged in shipment.
Service for Nikon Esclipse E400
You can locate an Authorized Nikon Repair Center for India at the following link:
http://www.nikon-asia.com/distributors.php
Eppendorf centrifuge model 5702: lid not does latch/lock after emergency release.
If the lid of your Eppendorf centrifuge model 5702 is not latching or locking closed after the emergency release, it is possible that the locking mechanism is damaged or misaligned. Here are some steps you can try to troubleshoot the issue:
- Check the alignment of the lid: Make sure that the lid is properly aligned with the base of the centrifuge. If it is not, the locking mechanism may not engage correctly. Adjust the lid as needed and try to latch it closed.
- Inspect the locking mechanism: Check the locking mechanism on both the lid and the base of the centrifuge for any signs of damage or wear. If you see any damage, you may need to replace the locking mechanism.
- Check the safety interlock switch: The safety interlock switch is a small button located on the base of the centrifuge that is depressed when the lid is closed. If this switch is not working correctly, it can prevent the lid from latching closed. Check that the switch is functioning properly by pressing it with a small tool while attempting to latch the lid.
- Contact Eppendorf customer support: If none of the above steps work, you may need to contact Eppendorf customer support for further assistance. They can help you troubleshoot the issue and provide advice on how to proceed.
3/14/2023 7:09:04 PM •
Science &...
•
194 views • 0 helpful votes
AA-Spectrometer Unicam 939 Error
The error you are experiencing on your Unicam 939 AA-Spectrometer with the software reporting error 16 and 2, along with lamps 1,2,3,4, and 10 lighting up, could indicate a few different issues. Here are some steps you can take to try to fix the problem:
- Restart the spectrometer: Turn off the spectrometer and unplug it from the power source. Wait a few minutes, then plug it back in and turn it on again. Sometimes, a simple restart can fix issues with the instrument.
- Check the calibration: It is possible that the instrument needs to be calibrated. Follow the manufacturer's instructions for calibrating the spectrometer.
- Check the lamps: If multiple lamps are lighting up, it could be due to a problem with the lamp assembly. Check that the lamps are properly installed and functioning. Try replacing any lamps that appear damaged or worn.
- Check the power supply: Make sure the power supply is functioning properly by testing the voltage output with a multimeter. If the voltage output is incorrect, try replacing the power supply.
- Check the communication between the software and the spectrometer: Ensure that the software is properly communicating with the spectrometer. Check that the cables are properly connected and that the software settings are correct.
If these steps do not resolve the issue, it may be necessary to contact the manufacturer for additional support or to seek the assistance of a professional repair technician.